Procedure— Proceed as directed for Procedure under 'Single-steroid Assay 511' (see below), using a solvent system consisting of a mixture of benzene and alcohol (98:2), through the fourth sentence of the second paragraph under Procedure. Then centrifuge the tubes for 5 minutes, and determine the absorbances of the supernatants in 1-cm cells at the wavelength of maximum absorbance at about 315 nm, with a suitable spectrophotometer, against the blank. [NOTE—Use 0.01 N alcoholic sodium hydroxide, rather than alcohol, to elute the silica gel bands.] Calculate the quantity, in mg, of C21H32O3 in the portion of Oxymetholone taken by the formula:
10C(AU / AS), in which C is the concentration, in mg per mL, of USP Oxymetholone RS (see below) in the Standard preparation, and AU and AS are the absorbances of the solutions from the Assay preparation and the Standard preparation, respectively.
single-steroid assay 511 - In the following procedure, the steroid to be assayed is separated from related foreign steroids and excipients by thin-layer chromatography and determined following recovery from the chromatogram.
Preparation of the Plate— Prepare a slurry from 30 g of chromatographic silica gel with a suitable fluorescing substance by the gradual addition, with mixing, of about 65 mL of a mixture of water and alcohol (5:2). Transfer the slurry to a clean, 20- × 20-cm plate, spread to make a uniform layer 250 µm thick, and allow to dry at room temperature for 15 minutes. Heat the plate at 105 for 1 hour, and store in a desiccator.
Solvent A— Mix methylene chloride with methanol (180:16).
Solvent B— Mix chloroform with acetone (4:1).
Standard Preparation— Dissolve in a mixture of equal volumes of chloroform and alcohol a suitable quantity of the USP Reference Standard specified in the individual monograph, previously dried as directed (see USP Reference Standards 11) and accurately weighed, to obtain a solution having a known concentration of about 2 mg per mL.
Assay Preparation— Prepare as directed in the individual monograph.
Procedure— Divide the area of the chromatographic plate into three equal sections, the left and right sections to be used for the Assay Preparation and the Standard Preparation, respectively, and the center section for the blank. Apply 200 µL each of the Assay Preparation and the Standard Preparation as streaks 2.5 cm from the bottom of the appropriate section of the plate. Dry the solution as it is being applied, with the aid of a stream of air. Using the Solvent specified in the individual monograph, develop the chromatogram in a suitable chamber, previously equilibrated and lined with absorbent paper, until the solvent front has moved 15 cm above the initial streaks.
Remove the plate, evaporate the solvent, and locate the principal band occupied by the Standard Preparation by viewing under UV light. Mark this band, as well as corresponding bands in the Assay Preparation and blank sections of the plate. Remove the silica gel from each band separately, either by scraping onto glazed weighing papers or by using a suitable vacuum collecting device, and transfer it to a glass-stoppered, 50-mL centrifuge tube. To each tube add 25.0 mL of alcohol, and shake for not less than 2 minutes. Centrifuge the tubes for 5 minutes, pipet 20 mL of the supernatant from each tube into a glass-stoppered, 50-mL conical flask, add 2.0 mL of a solution prepared by dissolving 50 mg of blue tetrazolium in 10 mL of methanol, and mix. Proceed as directed for Procedure under Assay for Steroids 351, beginning with “Then to each flask.”
USP Oxymetholone RS— Dry portion in vacuum over phosphorus pentoxide for 4 hours before using. Keep container tightly closed.
is this what i'm after ?